[Camino-users] DTI Data Quality and Possible Options (RESTORE)?

Michiko C michiko.rc at gmail.com
Tue Mar 17 18:34:41 PDT 2015


Hi Philip,

Thanks for the reply and for sharing your opinion regarding RESTORE.

Just wondering if you might be aware of any possible methods to get around
this issue? We have about 300 datasets and majority belong to longitudinal
studies, so we are seeking the best possible alternative to analyze the
data given this limitation.

Best wishes,
M

On Tue, Mar 17, 2015 at 5:38 AM, Philip A Cook <cookpa at mail.med.upenn.edu>
wrote:

> Hi,
>
> It does look like susceptibility artifact to me. See the example and
> discussion here
>
>
> http://onlinelibrary.wiley.com/doi/10.1002/nbm.1543/abstract;jsessionid=3118D8814A2571A49A79245DF1012B23.f01t02
>
> I'm not sure that RESTORE would be very helpful here since all the
> measurements in the affected region might be unreliable. RESTORE would be
> most helpful in the case where a few DWI measurements are detectable as
> outliers / bad data.
>
> On Mar 11, 2015, at 4:43 AM, Rowena Chin <michiko.rc at gmail.com> wrote:
>
> > Dear All,
> >
> > Would like to get some opinions and advice from helpful contributors on
> this forum regarding the quality of the DTI data that I have been asked to
> process. I have attached some screenshots in this post so it illustrates my
> question.
> >
> > 1) The datasets seem to have a similar problem which I'm guessing, is
> due to susceptibility artefacts. It seems to be a recurrent issue around
> the same areas and slices (around x=125-130, y=119-122, z=8-10).
> >
> > Based on the images, what would be the ideal way forward? Would using
> the RESTORE function help in this case?
> >
> > 2) The following parameters were used:
> >
> > - 3 T Philips scanner
> > - 15 directions in the same axial plane using single-shot, spin-echo
> echo planar imaging (EPI) sequence, b value of 0 and 800 s/mm2, TR/TE
> should be 3700/56, matrix 128 × 128, slice thickness 3 mm with no gap and
> FOV of 240 mm2.
> >
> > Based on the current parameters, is anyone able to provide advice on
> what could be altered to improve data quality? I am definitely not an
> expert when it comes to the MR physics - though I hope to understand it
> better. Could it be due to the 3mm slice thickness and through-slice
> dephasing?
> >
> > Best wishes,
> > Michiko
> >
> > <Screen Shot 2015-02-26 at 10.49.46 am.png><Screen Shot 2015-02-26 at
> 10.39.48 am.png>_______________________________________________
> > Camino-users mailing list
> > Camino-users at www.nitrc.org
> > http://www.nitrc.org/mailman/listinfo/camino-users
>
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://www.nitrc.org/pipermail/camino-users/attachments/20150318/ea4fc935/attachment.html>


More information about the Camino-users mailing list